[Histonet] Knife angle, old 820 disposable holder and block a
ngle?further comments
DiCarlo, Margaret
MDiCarlo <@t> KaleidaHealth.Org
Fri May 7 13:13:23 CDT 2004
I use Accu-Edge high profile blades and cut at 4 microns and have great
sections using the non adjustable knife holder for the old AO microtomes.
Peggy DiCarlo
-----Original Message-----
From: S Ladd [mailto:sladd <@t> hsc.usf.edu]
Sent: Friday, May 07, 2004 13:42
To: Histonet <@t> lists.utsouthwestern.edu
Subject: RE: [Histonet] Knife angle, old 820 disposable holder and block
angle?further comments
I have an old AO 820 microtome and I have the non-adjustable blade holder. I
use DuraEdge high profile blades (#7310)--1-800-253-2768 and they work great
(even with 4 micron sections)!! I couldn't get the Leica blades to work at
all.
I have also tried an Reichert insert-type holder in my adjustable steel
knife holder but that didn't seem to work as well either.
Again, a lot of this is just personal preference. I would try some different
types of blades before you give up on the non-adjustable blade holder.
Sharron
-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu]On Behalf Of Gayle
Callis
Sent: Friday, May 07, 2004 1:02 PM
To: KAELPERS <@t> aol.com; Histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Knife angle, old 820 disposable holder and block
angle?further comments
Felt the need for further comment about blade angle with older AO 820
microtome.
You wrote: "It is correct that the blade holder is permanent and the only
way to adjust the angle is with the block holder".
You do not adjust the blade angle i.e the cutting angle at the block holder
- block holder adjustment (the x/y axis adjusment) only provides is
orientation of block face to the blade which is set at its cutting angle.
Once block is trimmed and ready for sectioning, you are locked into using
the cutting angle - determined by blade holder and bevel angle ON
disposable blade. With today's tissue cassettes, block face adjustment has
become blessedly minimal - we only readjust block face for some block
recuts but we do change holder adjustments to accomodate the blade being
used.
We found using old AO 820 blade holder that some brands of knives are
slightly different due to how they were manufacturered/treated. Some
blades just work better than others in this holder, and you may have have
to try various brands. Ask for samples if you experience too much
difficulty with any one disposable. blade.
A significant reason why blade holders are now totally adjustable is
probably because the old holder WAS a permanent angle. Microtomists still
need to refine cutting angles for any given blade use. There have been long
discussions about this on Histonet, check out archives. One can use the
820 standard knife holder (for C profile steel knives), buy a disposable
blade insert to achieve fine tweaking of cutting angle. This may be
important IF one really dislikes or cannot work with old 820 holder.
Have a good weekend after this long (whew!) lecture
Gayle Callis
MT,HT,HTL(ASCP)
Research Histopathology Supervisor
Veterinary Molecular Biology
Montana State University - Bozeman
PO Box 173610
Bozeman MT 59717-3610
406 994-6367 (lab with voice mail)
406 994-4303 (FAX)
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
CONFIDENTIALITY NOTICE:
This email transmission and any documents, files,
or previous e-mail messages attached to it are
confidential and intended solely for the use of the
individual or entity to whom they are addressed.
If you are not the intended recipient, or a person
responsible for delivering it to the intended recipient,
you are hereby notified that any further review,
disclosure, copying, dissemination, distribution, or
use of any of the information contained in or attached
to this e-mail transmission is strictly prohibited.
If you have received this message in error, please
notify the sender immediately by e-mail, discard
any paper copies, and delete all electronic files
of the message. If you are unable to contact the
sender or you are not sure as to whether you
are the intended recipient, please e-mail
ISTSEC <@t> KaleidaHealth.org or call (716) 859-7777.
More information about the Histonet
mailing list