Rita van Bendegem van <@t> uhnres.utoronto.ca
Thu Oct 2 08:04:12 CDT 2003


Dear All,

I have had some problems with Superfrosted slides but in our research lab we
process our IHC slides manual.  We react both frozens and paraffin tissues. Our
routine chromogens that we use are DAB, VIP, Alexa 488, and Alexa 594,
depending on the tissue and  the application.  We have been using Fisherbrand
Superfrosted Plus slides for years with no known problems.  In February of this
year I discovered that NF200 antibody from Sigma was not working at all in
frozen or paraffin tissue, but my positive control tissue was O.K. (since it was
cut on to older Superfrosted slides)
I pulled out some old paraffin slides dating back to Jan. 2002 and Sept. 2002
then ran the NF200 immuno.  The Jan. 2002 slides were positive, Sept. 2002 were
weakly positive and I also ran another Feb. 2003 slide which was negative.

This made me call my Fisher rep. about the problems and he in return called the
manufacture of the Superfrosted Plus slides, which from my understanding is ERIE
in North America. The Fisher rep. told me that he never got a response back from
the manufacture as of a few months ago.

In the meantime, I called a few techs  and made some inquires.  I was told to
try some slides from Surgipath that are called X-tra Adhesive they are European
slides with  USA, Canada, France and UK address written on the side of box.   I
have used them and have had no problems with my NF200 controls since.

I would be interested to hear if anyone else has had similar problems .




Bill Sinai wrote:

> Dear All,
>
> Has anyone had a problem with Superfrost Plus slides on automated
> Immunohistochemistry Staining machines?  If so what have the problems been?
> Also if the problem was resolved and how it was resolved.
>
> We obtain our slides through a third party agency in Australia.  We are at
> present using Menzel-Glaser (German) brand.  I am led to believe they are
> the largest slide and coverglass manufacturer in the world.
>
> Bill Sinai
> Laboratory Manager
> Tissue Pathology, ICPMR
> Westmead NSW 2145
> Australia
> Ph 02 9845 7774
>
> __________________________________________________________________
>
> This electronic message and any attachments may be confidential.  If you
> are not the intended recipient of this message would you please delete the
> message and any attachments and advise the sender. Western Sydney
> Area Health Services (WSAHS) uses virus scanning software but excludes
> any liability for viruses contained in any email or attachment.
>
> This email may contain privileged and confidential information intended
> only for the use of the addressees named above. If you are not the
> intended recipient of this email, you are hereby notified that any use,
> dissemination, distribution, or reproduction of this email is prohibited. If
> you have received this email in error, please notify WSAHS
> immediately.
>
> Any views expressed in this email are those of the individual sender
> except where the sender expressly and with authority states them
> to be the views of WSAHS.
>
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet

--
Rita van Bendegem, MLT
Laboratory Manager,
Canadian Paraplegic Association,
Spinal Cord Injury Research Centre,
Mc Laughlin Pavilion Lab 12-405,
399 Bathurst St.,
Toronto Western Hospital,
University Health Network,
Toronto, Ontario  M5T 2S8

Telephone 416-603-5800 ext. 2792
Fax 416-603-5745






More information about the Histonet mailing list