[Histonet] IHC background
    Regan Fulton 
    regan.fulton at gmail.com
       
    Thu Sep 29 14:42:11 CDT 2022
    
    
  
Hi,
The artifact you are seeing is a well-known artifact of the Leica covertile
design.
It does limit the available footprint for placing samples.
Hope this helps.
Regan
Regan Fulton, M.D., Ph.D.
CEO and Co-Founder
Array Science, LLC
475 Gate 5 Road, #100
Sausalito, CA 94965
(415) 577-7360
www.arrayscience.com
On Thu, Sep 29, 2022 at 12:37 PM RENE MORALES via Histonet <
histonet at lists.utsouthwestern.edu> wrote:
> Paula,
> Regardless of the position of the tissue: control/ patient ‘s tissue, most
> likely is the  distribution of the reagents during staining, isn’t
> adequate. Syringe, staining  reagents expiration, calibration, etc.,   Have
> you contacted service from Leica? Hope the solution is found,
> Thanks,
>
>
> Sent from my iPhone
>
> > On Sep 29, 2022, at 12:00, Paula via Histonet <
> histonet at lists.utsouthwestern.edu> wrote:
> >
> > Hello everyone,
> >
> >
> >
> > We have the Bond III and we ran an S100.  We placed 2 serial sections
> > vertically on the slide with the control tissue at the top for staining.
> The
> > patient's section at the top underneath the control tissue had background
> > staining, while the section on the bottom of that had no background
> > staining.  Can someone determine why this would happen?
> >
> >
> >
> > Thank you in advance.
> >
> > Paula
> >
> > Bio-Path Medical Group
> >
> > _______________________________________________
> > Histonet mailing list
> > Histonet at lists.utsouthwestern.edu
> > http://lists.utsouthwestern.edu/mailman/listinfo/histonet
>
>
> _______________________________________________
> Histonet mailing list
> Histonet at lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
>
    
    
More information about the Histonet
mailing list