[Histonet] RE: ThinPrep
Yang, Mari
MYang <@t> emc.org
Mon Sep 16 11:30:52 CDT 2013
Lisa,
If the ThinPrep processor is new to your organization, then the requirements below may apply.
CYP.05257
Implementation/Verification Protocol
Phase II
There is documentation of adherence to the manufacturer's recommended protocol(s) for implementation and verification of new instruments.
NOTE: Before implementing use of new gynecologic liquid-based methods and instruments, automated preparations, and automated screening instruments, the laboratory must verify and document the functioning of the instrument in its own specific laboratory environment, including the capability of the instrument to replace existing procedure(s), if applicable. If the manufacturer does not provide verification and instrument monitoring recommendations, the laboratory must document the specific verification procedure used.
Evidence of Compliance:
✓ Records of completed instrument/method verification consistent with manufacturer's recommendations OR records of an alternative documented verification procedure approved by the section director
REFERENCES
1)
Chang AR, et al. Can technology expedite the cervical cancer screening process? A Hong Kong experience using the AutoPap Primary Screening System with location-guided screening capability. Am J Clin Pathol. 2002; 117:399-437-443
2)
Wilbur DC, et al. Location-guided screening of liquid-based cervical cytology specimens. A potential improvement in accuracy and productivity is demonstrated in a preclinical feasibility trial. Am J Clin Pathol. 2002; 118:399-407
3)
Department of Health and Human Services, Centers for Medicare and Medicaid Services. Clinical laboratory improvement amendments of 1988; final rule. Fed Register. 2003(Jan 24):3707 [42 CFR493.1253, 493.1255, and 493.1274(g)]
Good Luck!
Mari
Mari Yang, MHA, CT(ASCP)CMHTLCM
Cytology Supervisor
Eisenhower Medical Center
Tel: 760.773.2009
-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Joe W. Walker, Jr.
Sent: Saturday, September 14, 2013 11:45 AM
To: White, Lisa M.; histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] RE: ThinPrep
Hi Lisa,
I am not aware of a specific CAP regulation/question related to the validation of the ThinPrep processor. However, there are general questions related to ensuring your devices are functioning as intended prior to using them on patient specimens. This should have been done when the device was initially setup in the lab and with the help of the manufacturer. They should help you verify that the instrument is working as designed.
Joe W. Walker, Jr. MS, SCT(ASCP)CM
Anatomical Pathology Manager
Rutland Regional Medical Center
160 Allen Street, Rutland, VT 05701
P: 802.747.1790 F: 802.747.6525
Email joewalker <@t> rrmc.org www.rrmc.org
Our Vision:
To be the Best Community Healthcare System in New England
Rutland Regional...Vermont's 1st Hospital to Achieve Both ANCC Magnet Recognition® and the Governor's Award for Performance Excellence
-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of White, Lisa M.
Sent: Thursday, September 12, 2013 9:16 AM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] ThinPrep
Does anyone know a CAP regulation for validation of ThinPrep?
Lisa White HT(ASCP)
Supervisory HT
James H. Quillen VAMC
Corner of Veterans Way and Lamont
VAMC Warehouse BLDG. 205
PO Box 4000
PLMS 113
Mountain Home, TN 37684
423-979-3567
423-979-3401 fax
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
This message (and any included attachments) is from Rutland Regional Health Services and is intended only for the addressee(s). The information contained herein may include privileged or otherwise confidential information. Unauthorized review, forwarding, printing, copying, distributing, or using such information is strictly prohibited and may be unlawful. If you received this message in error, or have reason to believe you are not authorized to receive it, please promptly delete this message and notify the sender by e-mail.
Thank You
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
More information about the Histonet
mailing list