[Histonet] Reagent rotation

Goins, Tresa TGoins <@t> mt.gov
Tue Jun 5 13:35:14 CDT 2012


We rotate by cassette number - our daily load can vary from 10 to more than 100 cassettes, so an exchange schedule based on number of runs was not appropriate.

Tresa

-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Mark Perrin
Sent: Tuesday, June 05, 2012 12:05 PM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Reagent rotation

What is the normal procedure for changing the reagents in the tissue processor? We have a Sakura VIP6. Should the schedule of changing be based on number of runs vs. number of blocks processed? About 80% of our blocks are small GI biospies.

Thanks in advance for your input.

 
Toshia Perrin