[Histonet] fresh-frozen mouse tail sectioning
Patsy Ruegg
pruegg <@t> ihctech.net
Fri May 16 11:53:40 CDT 2008
I would just lay it on it's side for freezing the first time and then turn
it vertically to mount onto a chuck for sectioning. Also, you probably do
not have this, but for difficult things like this (the tail will have at
least cartilage in it) I use the tape transfer technique from Instrumedics
(now owned by McCormick).
Patsy
Patsy Ruegg, HT(ASCP)QIHC
IHCtech, LLC
Fitzsimmons BioScience Park
12635 Montview Blvd. Suite 215
Aurora, CO 80010
P-720-859-4060
F-720-859-4110
wk email pruegg <@t> ihctech.net
web site www.ihctech.net
This email is confidential and intended solely for the use of the Person(s)
('the intended recipient') to whom it was addressed. Any views or opinions
presented are solely those of the author. It may contain information that is
privileged & confidential within the meaning of applicable law. Accordingly
any dissemination, distribution, copying, or other use of this message, or
any of its contents, by any person other than the intended recipient may
constitute a breach of civil or criminal law and is strictly prohibited. If
you are NOT the intended recipient please contact the sender and dispose of
this e-mail as soon as possible.
-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Gaupp, Dina
D
Sent: Friday, May 16, 2008 10:05 AM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] fresh-frozen mouse tail sectioning
To Whomever This Applies:
I am having big problems sectioning mouse tail, approximately 2mm in
thickness, vertical embedded in OCT for frozen sections. The tissue is
fresh-frozen because the principle investigator would like to detect an
enzyme x-gal. No fix, no cryopreservation - the tissue was not immersed in
any type of solution. He snipped the ends of a mouse tail & immediately
gave me the tissues. I embedded the tissue vertically in OCT & flash froze
at -80C. Upon sectioning, the tissue rolled. I could not for the life of
me get 1 section. All the tissues rolled & its so tiny to begin with that
it was hard for me to grab it. I used the anti-roll plate, hoping it would
hold the tissue in place but it still rolled. I changed temperature
settings(increase & decrease temp), thickness, angle, rubbed it with my
fingers. Everything I could think of to get a section. I asked someone
else in the lab to try & they couldn't get a section. It was like the
tissue completely separated from OCT. The principle investigator will give
me more samples & I don't want this to happen again.
Can anyone in histoland help me, tell me what I didn't do or what I did
wrong?
Dina D. Gaupp, BS, MT
Senior Lab Supervisor
Center for Gene Therapy, SL-99
Tulane University Health Science Center
1430 Tulane Ave
New Orleans, La 70112
Lab: 504-988-1194
dgaupp <@t> tulane.edu <mailto:dgaupp <@t> tulane.edu>
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
More information about the Histonet
mailing list