[Histonet] Cryostat sections with small bubbles
Alan Bright
abright <@t> brightinstruments.com
Fri Jun 17 04:10:09 CDT 2005
Dear Kathleen,
Thank you for your reply. I do understand that Specimen Temperature Control may not be available to all cryostat users. Therefore I will set out a method that I have posted on Histonet previously to assist with sectioning problems on brain. You will see that this is a far more scientific method than warming up the face of a specimen with ones finger and will render repeatable good quality sections.
Set the cryostat microtome chamber to -8n to -12ºC, place containers made from aluminium kitchen foil around and in contact with the knife in areas that with not impede the specimens travel or the anti-roll plate, also do the same for where the anti-roll plate is parked away from the knife. Fill these containers with graduals of Solid C02. Allow 10 minutes for the solid C02 to cool the knife/ anti-roll plate to a lower temperature than the specimen, then start sectioning.
Best Regards
Alan Bright
Bright Instrument Co.Ltd.
St Margaret's Way
Huntingdon
Cambridgeshire
PE29 6EU
England
Tel No:+44 (0)1480 454528
Fax No:+44 (0)1480 456031
Email: abright <@t> brightinstruments.com
Web Site: www.brightinstruments.com
-----Original Message-----
From: Kathleen Spencer [mailto:kspencer <@t> utmem.edu]
Sent: 16 June 2005 16:05
To: Alan Bright
Cc: Emily Jane Wiesner-Camm; histonet <@t> lists.utsouthwestern.edu; Katri Tuomala
Subject: Re: [Histonet] Cryostat sections with small bubbles
Yes Alan, but when I put my finger on the block face to warm it, there is a kimwipe or gauze between my finger and the tissue. It works really well, so I guess the Ark is still around.
As the owner of a company that sells cryostats, you need to be aware that not everyone has or can obtain such devices.
Many of us, especially in research, work with old and crippled equipment, thus we need all the help we can get from our fellow technicians.
Kathleen Spencer, HT ASCP
Lab Manager/LCM Supervisor
UTHSC
On Jun 16, 2005, at 5:11 AM, Alan Bright wrote:
Emily,
Basically you need to section brain at -8 to -12ºC as colder temperatures will cause what you are getting plus cracking. However to achieve good quality sections at these temperatures you will need to maintain the knife and anti-roll temperatures at -20ºC or lower to allow the section to slide in between the knife and the anti-roll plate unimpeded. Which is what Kathleen is doing when she warms the block face with her finger, a technique that went out with the Ark and should not be used as firstly you could have an infection issue and secondly it is not a very scientific approach in this day and age when there are devices fitted to some cryostats such as independent specimen temperature control to perform this task with full and repeatable control and of most importance, ease.
Best Regards
Alan Bright
Bright Instrument Co.Ltd.
St Margaret's Way
Huntingdon
Cambridgeshire
PE29 6EU
England
Tel No:+44 (0)1480 454528
Fax No:+44 (0)1480 456031
Email: abright <@t> brightinstruments.com
Web Site: www.brightinstruments.com
-----Original Message-----
From: Kathleen Spencer [mailto:kspencer <@t> utmem.edu]
Sent: 15 June 2005 16:50
To: Katri Tuomala
Cc: histonet <@t> lists.utsouthwestern.edu; Emily Jane Wiesner-Camm
Subject: Re: [Histonet] Cryostat sections with small bubbles
Hi Emily,
I cut rat brain sections at -18 to -20 and often have to warm the block
face with my finger before each section. I also have better luck if I
cut sections at 10u. These are fresh frozen brains. If I cut them at
20u I get bubbles under the sections. I keep them in the cryostat, and
then fix the sections in cold fix, buffer wash, water wash, then air
dry them in the hood.
When I cut rat brain that has been perfused, I cut 20u floating
sections. They never adhere well to slides, even if I use a hot plate.
I always have bubbles under the section. That is why in this case we
use floating sections.
I hope this is helpful.
Kathleen Spencer
On Jun 14, 2005, at 8:39 PM, Katri Tuomala wrote:
Hi Emily,
I have never heard of hot plating cryostat sections. You may be
boiling the moisture under the section forming bubbles or do you air
dry them well first? Also -35 C seems too cold for brain sections, you
may be getting some cutting artifacts.
I'm sure there'll be others with helpful information for you.
Katri
Katri Tuomala
Hamilton, Ontario, Canada
----- Original Message ----- From: "Emily Jane Wiesner-Camm"
<Emily.Wiesner <@t> medecine.unige.ch>
To: <histonet <@t> lists.utsouthwestern.edu>
Sent: Tuesday, June 14, 2005 10:36 AM
Subject: [Histonet] Cryostat sections with small bubbles
Hi All,
I was wondering whether anyone can let me know why small bubbles
appear underneath my rat brain sections when I cut them on a cryostat
(at -35) when they appear perfectly flat before placing them on a
hotplate. How they can be avoided?
Thanks,
Emily
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu http://lists.utsouthwestern.edu/mailman/listinfo/histonet
More information about the Histonet
mailing list